● DISCUSSION ofDialyzer Reuse
INTRODUCTION
Dialysis facilities that reuse dialyzers must follow strict guidelines set forth by the AAMI
(Association for the Advancement of Medical Instrumentation)
The AAMI guidelines make provisions for the patient’s safety, as listed below:
● A dialyzer must be clearly labelled with the patient’s name and should only be
used for the same patient.
● A dialyzer must be tested after each use to make sure it is working properly.
● A dialyzer must be tested after rinsing for any traces of disinfectant that may
remain.
Patients must be monitored for any reactions due to reuse.
REQUIREMENTSFORREPROCESSINGDIALYZERS
WATER
● This should be of AAMI standard or preferably of ultrapure quality.
● The water line pressure should be 13 kg/cm2 (20 psi), and every washing area
should be equipped with 2 outlets or a T connection.
● Two different fittings will be required on the water line at each reprocessing area,
a standard tubing to clean the blood compartment, and a Hansen connector for
backwashing the dialysate compartment.
● Fittings should be of 316 SS or medical grade PVC only
A prolonged rinse with Reverse Osmosis (RO) water followed by reverse ultra-
filtration has been shown to be an effective means of cleaning dialyzers without
the addition of chemicals
CLEANING AND DISINFECTING AGENTS
● While selecting disinfectant we need to see for – Elimination of organic pollution,
Inorganic component removal capacity, Disinfection efficacy, Safety and
Environmental pollution.
These agents should preferably be available online in the reprocessing areas.
● Overhead tanks containing the chemicals may be of 25 to 50 liters capacity and
should be 
 refilled with fresh solutions every week, after cleaning.
● All tanks and piping for sodium hypochlorite should be composed of medical
grade PVC, and those for formaldehyde, glutaraldehyde, and peracetic acid
should be composed of 316SS.
● As per Indian Haemodialysis guidelines, common cleansing/disinfecting agents in
use are mentioned below but regardless of the type of cleaning or disinfecting
agent used in the facility, a thorough water rinse must be done prior to adding
chemicals and after cleaning and disinfecting. Safety tests must be done after the
final rinse to confirm the absence of any chemical used.
● It is dangerous to rely on a timed rinse without the use of a valid safety test.
Residual testing prevents patient injury due to chemical exposure.
a) Sodium Hypochlorite1–2%
● Sodium hypochlorite (bleach) is a cold disinfectant.
Commercially available cans are 10% and they need to be diluted to the above
concentration.
● There is formation of free chlorine that is a strong oxidant. It cleans and
eliminates cellular debris.
● The advantages are that it is cheap, effective and safe.
Residual testing for sodium hypochlorite is done with chlorine reagent test strips
with a limit of 0.5 parts per million (ppm). It is a very sensitive method. Sodium
hypochlorite in minute amounts greater than 1:25,000 produces hemolysis.
● Failure to perform residual testing will result in acute hemolysis or slow hemolysis
that may go undetected.
● Sodium hypochlorite (bleach), diluted to 0.6% or less, dissolves proteinaceous
deposits that may occlude fibers. Bleach increases albumin losses in high-flux
cellulose triacetate (CT 190) and polysulfone.
● Polyvinylpyrrolidone dialyzers Albumin losses are generally not clinically
significant unless bleach is used on high-flux membranes with exceptionally high
water permeability.
b) Hydrogen Peroxide
● Used as a cleaning agent.
● 1% or less concentration.
● It does not remove protein deposits.
● RO water should be used for dilution.
● Must be stored in airtight container as it is unstable.
● This is instilled in the dialysate compartment of the dialyzer rather than the blood
compartment.
● Though hydrogen peroxide has been shown to have a lower potency than
hypochlorite, 2 Indian studies showed good results after reuse with this agent.
● Test strips detect peroxide residues with limit of 0.5mg/l.
● Range: 0-0.5-2.5-5-10-25mg/l H2O2.
● Colour change: white > blue.
● Residues should be NIL.
c) Formaldehyde (4%)
● It is the most common disinfectant used for fluid delivery systems.
● Formaldehyde has no cleaning properties. Formaldehyde denatures protein and
fixes most cellular debris.
● Formaldehyde is a cold sterilant that effectively kills all microorganisms, including
spores and resistant viruses but with risk of serious septicemia, if inadequate
formaldehyde concentrations are being used.
● Commercially available as 40%, this can be diluted with the water used for
reprocessing to give a final strength of 4%.
● Concentrations lower than 4% formaldehyde do not adequately kill
Mycobacterium chelonae in water
● It is an inexpensive and stable solution with a long shelf-life. Formaldehyde is a
gas that is dissolved in water to form the compound formalin.
● The formaldehyde gas is irritating to the eyes and has an offensive odour. Gloves
must always be worn when handling formaldehyde to prevent dermatitis and
allergic sensitivities. The room must be well ventilated. Any splashing must be
minimized. A face shield gives total protection to the face. Minimally, eye
protection (goggles) must be worn when handling formaldehyde.
● Sensitive residual testing is now available with limit of 1.0 ppm. The same
principles apply as described for peracetic acid. In the past, safety tests for
formaldehyde used Schiff’s reagent. Schiff’s reagent will test to 5 ppm. The
newer indicator test strips are more sensitive.
d) Glutaraldehyde (2%)
● This has to be freshly prepared and activated.
● The chemical potency of the solution may be tested with Schiff’s reagent which

 produces a magenta colour similar to that seen with formaldehyde.
● Glutaraldehyde should be used for preserving the end caps, universal connectors, O-
rings, and dialyzer caps when not in use.
● The solution should be replaced at intervals of not less than 10 days.
● Small containers containing glutaraldehyde should be available both at the dialysis

 stations and at the reprocessing areas.
e) PeraceticAcid(Renalin/Hemoclean, etc.)
Hemoclean/Renalin
● A mixture of peracetic acid, hydrogen peroxide, & acetic acid
○ PERACETIC ACID: Disinfection during short term.
○ HYDROGEN PEROXIDE: Disinfection during long terms, Isolate protein
pollutants.
○ ACETIC ACID: Cleaning effect (Remove inorganic pollutants)
Strong bactericidal, sporicidal, fungicidal, & virucidal.
● It is probably the cold sterilant of choice for fluid delivery systems. Unlike formaldehyde,
this mixture leaves no toxic residues. It decomposes into oxygen and acetic acid after
reacting with organic material.
● The mixture is a strong oxidant that readily cleans all cellular debris, precipitates, and
scale in the machines when used routinely.
● Minimum concentration 1%. It can be used as cleaning agent & sterilant. The undiluted
solution should be diluted to prepare 2 solutions of the following:
●
2% (200 ml in 10 liters of water) as a cleaning agent and
3.5% (350 ml in 10 liters of water) as a disinfectant
● Minimum dwell time 11 hours.
● Stable for 1 week.
● Avoid direct sunlight exposure of solution.
● The odour is pungent.
● Test strips are available to check for the presence or absence of the peracetic acid
mixture. Indicator test strips ensure the adequate presence of this cold sterilant in the
machines and residual test strips test confirms absence.
● Range of indicator test strips: 0-500-1000-1500-2000 mg/l peracetic acid.
● Colour change: bright yellow > red.
If the sterilant inside the dialyser shows less than 500 PPM it should not be used ideally
as it suggests there is inadequate strilent effect. Now some other Countries specially
Japan has increased this limit to 800 PPM.
● Once the confirmation is done that the potency is more than 500 or 800 PPM then prime
the dialyser with NS.
● After the peracetic acid mixture is rinsed from the machine, residual test strips test for the
absence of the peracetic acid mixture. The dialyzer should be primed with at least 2000
ml of 0.9% normal saline using the dialysis machine blood pump at a speed of 150 ml/
min. The dialysate lines should be connected and the dialysate compartment filled with
dialysate flowing at 500 ml/ min prior to starting the priming procedure. Failure to “dialyze”
the disinfectant out may result in inadequate removal and reactions after starting dialysis.
● After 2000 ml of saline priming, effluent from the venous line should be checked for the
presence of residual disinfectant.
● The residual testing is very sensitive with limit <1 ppm.
● Because the peracetic acid mixture is a strong oxidant, its handling requires careful
attention to avoid chemical burns. Gloves and face protection are mandatory. Accidental
contact exposure to this chemical requires water flushes and medical attention similar to
formaldehyde.
REPROCESSING OF TUBINGS 

● No literature supports the reuse of blood tubings, however, they are reused in most units.
● The tubings are washed free of blood by treated water of AAMI or EU standard, and then
with a 1.6% solution of sodium hypochlorite.
● The arterial and venous bubble chambers are gently tapped to release clots, and the side
tubings are all cleared by clamping the outlets to dislodge any adherent material.
● The tubings are again rinsed with water and then connected to a supply of 4%
formaldehyde, which is allowed to completely displace water and air from the tubings.
TESTS OF PERFORMANCE
● No objective tests of performance are available for blood tubings.
● The tubings are discarded if the normal elasticity appears to be lost, if there are visible
cracks, if there is a change from the normal transparent appearance, or damage to any of
the hubs.
● A test that is sometimes useful is to compare the elasticity of the pump segment with that
of new tubing on a blood pump.
● Failure to give 90% of the flow obtained with a new tubing segment or “slipping” of the
tubing or a “slapping sound” from the rollers may indicate a malfunction of the pump
segment of the tubing and require it to be discarded.
Mention By-
Anup Singh
Dialysis Technician and Trainer
9651633399

More Related Content

What's hot

Hemodynamic monitoring for hemodialysis pt
Hemodynamic monitoring for hemodialysis ptHemodynamic monitoring for hemodialysis pt
Hemodynamic monitoring for hemodialysis ptlisabaizura
 
ADEQUACY OF HEMODIALYSIS
ADEQUACY OF HEMODIALYSISADEQUACY OF HEMODIALYSIS
ADEQUACY OF HEMODIALYSISsaihari17
 
Dialysis without anticoagulation (Heparin Free Dialysis)
Dialysis without anticoagulation (Heparin Free Dialysis)Dialysis without anticoagulation (Heparin Free Dialysis)
Dialysis without anticoagulation (Heparin Free Dialysis)Mahmoud Eid
 
Peritoneal dialysis part1
Peritoneal dialysis part1Peritoneal dialysis part1
Peritoneal dialysis part1FarragBahbah
 
Anticoagulation in hemodialysis
Anticoagulation in hemodialysisAnticoagulation in hemodialysis
Anticoagulation in hemodialysisVishal Ramteke
 
Basics of peritoneal dialysis
Basics of peritoneal dialysisBasics of peritoneal dialysis
Basics of peritoneal dialysisVishal Golay
 
Dialysis machines.pptx
Dialysis machines.pptxDialysis machines.pptx
Dialysis machines.pptxDarshanS239776
 
vascular access for dialysis access: seminar
vascular access for dialysis access: seminarvascular access for dialysis access: seminar
vascular access for dialysis access: seminarMd Rahman
 
Hemodialysis vascular catheters review
Hemodialysis vascular catheters review Hemodialysis vascular catheters review
Hemodialysis vascular catheters review JAFAR ALSAID
 
Extracorporeal Therapy - Dr. Samir kamal
Extracorporeal Therapy - Dr. Samir kamalExtracorporeal Therapy - Dr. Samir kamal
Extracorporeal Therapy - Dr. Samir kamalMNDU net
 
DIALYSIS - Access, Hemo dialysis
DIALYSIS -   Access, Hemo dialysis DIALYSIS -   Access, Hemo dialysis
DIALYSIS - Access, Hemo dialysis Shanta Peter
 

What's hot (20)

Hemodynamic monitoring for hemodialysis pt
Hemodynamic monitoring for hemodialysis ptHemodynamic monitoring for hemodialysis pt
Hemodynamic monitoring for hemodialysis pt
 
Permnent vascular access
Permnent vascular accessPermnent vascular access
Permnent vascular access
 
ADEQUACY OF HEMODIALYSIS
ADEQUACY OF HEMODIALYSISADEQUACY OF HEMODIALYSIS
ADEQUACY OF HEMODIALYSIS
 
Sled 2019
Sled 2019Sled 2019
Sled 2019
 
History of dialysis
History of dialysisHistory of dialysis
History of dialysis
 
Dialysis without anticoagulation (Heparin Free Dialysis)
Dialysis without anticoagulation (Heparin Free Dialysis)Dialysis without anticoagulation (Heparin Free Dialysis)
Dialysis without anticoagulation (Heparin Free Dialysis)
 
Dialysis basics
Dialysis basicsDialysis basics
Dialysis basics
 
Peritoneal dialysis catheter
Peritoneal dialysis catheterPeritoneal dialysis catheter
Peritoneal dialysis catheter
 
Dialysis Water Treatment
 Dialysis Water Treatment  Dialysis Water Treatment
Dialysis Water Treatment
 
Peritoneal dialysis part1
Peritoneal dialysis part1Peritoneal dialysis part1
Peritoneal dialysis part1
 
Anticoagulation in hemodialysis
Anticoagulation in hemodialysisAnticoagulation in hemodialysis
Anticoagulation in hemodialysis
 
Basics of peritoneal dialysis
Basics of peritoneal dialysisBasics of peritoneal dialysis
Basics of peritoneal dialysis
 
Dialysate
DialysateDialysate
Dialysate
 
Dialysis machines.pptx
Dialysis machines.pptxDialysis machines.pptx
Dialysis machines.pptx
 
vascular access for dialysis access: seminar
vascular access for dialysis access: seminarvascular access for dialysis access: seminar
vascular access for dialysis access: seminar
 
Anticoagulation
AnticoagulationAnticoagulation
Anticoagulation
 
Hemodialysis vascular catheters review
Hemodialysis vascular catheters review Hemodialysis vascular catheters review
Hemodialysis vascular catheters review
 
Extracorporeal Therapy - Dr. Samir kamal
Extracorporeal Therapy - Dr. Samir kamalExtracorporeal Therapy - Dr. Samir kamal
Extracorporeal Therapy - Dr. Samir kamal
 
DIALYZER
DIALYZERDIALYZER
DIALYZER
 
DIALYSIS - Access, Hemo dialysis
DIALYSIS -   Access, Hemo dialysis DIALYSIS -   Access, Hemo dialysis
DIALYSIS - Access, Hemo dialysis
 

Similar to Discussion of dialyzer reuse

opthalmic preparations, Classification,factors affecting for the drug given t...
opthalmic preparations, Classification,factors affecting for the drug given t...opthalmic preparations, Classification,factors affecting for the drug given t...
opthalmic preparations, Classification,factors affecting for the drug given t...krishna keerthi
 
Scale up techniques in the production of sterile products &amp; ophthalmic pr...
Scale up techniques in the production of sterile products &amp; ophthalmic pr...Scale up techniques in the production of sterile products &amp; ophthalmic pr...
Scale up techniques in the production of sterile products &amp; ophthalmic pr...Dr. Jigar Vyas
 
Abdos Cataloge International
Abdos Cataloge InternationalAbdos Cataloge International
Abdos Cataloge InternationalShailendra Yadav
 
Compendial methods for evalution of crude drug and herbal formulation
Compendial methods for evalution of crude drug and herbal formulationCompendial methods for evalution of crude drug and herbal formulation
Compendial methods for evalution of crude drug and herbal formulationYachita Rajwadwala
 
Tlc for pesticide analysis copy
Tlc for pesticide analysis copyTlc for pesticide analysis copy
Tlc for pesticide analysis copychemnidhi
 
Practice Guidelines Endocopic disinfection & Reprocessing
Practice Guidelines Endocopic disinfection & ReprocessingPractice Guidelines Endocopic disinfection & Reprocessing
Practice Guidelines Endocopic disinfection & ReprocessingHossam Ghoneim
 
Chemical Oxygen Demand
Chemical Oxygen DemandChemical Oxygen Demand
Chemical Oxygen DemandMuhd Effandi
 
Chemical Methods for Sterilization
Chemical Methods for  SterilizationChemical Methods for  Sterilization
Chemical Methods for SterilizationPriyaDixit46
 
014.irrigation and intracanal medicaments
014.irrigation and intracanal  medicaments014.irrigation and intracanal  medicaments
014.irrigation and intracanal medicamentsDr.Jaffar Raza BDS
 
Miscellaneous test .pdf
Miscellaneous test .pdfMiscellaneous test .pdf
Miscellaneous test .pdfUVAS
 
Medicall genetics lab manual وراثة عملي
Medicall genetics lab manual وراثة عملي  Medicall genetics lab manual وراثة عملي
Medicall genetics lab manual وراثة عملي في رحاب الله
 
in process quality control test for ophthalmic and parenterals
in process quality control test for ophthalmic and parenterals in process quality control test for ophthalmic and parenterals
in process quality control test for ophthalmic and parenterals Henisha Patel
 
manufacturing of Parenterals.pdf
manufacturing of Parenterals.pdfmanufacturing of Parenterals.pdf
manufacturing of Parenterals.pdfSohailSheikh62
 
Human orosomucoid-2-elisa-kit
Human orosomucoid-2-elisa-kitHuman orosomucoid-2-elisa-kit
Human orosomucoid-2-elisa-kitDavid tong
 
Good Chromatographic Practices
Good Chromatographic PracticesGood Chromatographic Practices
Good Chromatographic PracticesDr. Amsavel A
 

Similar to Discussion of dialyzer reuse (20)

parenteral solution
 parenteral solution parenteral solution
parenteral solution
 
Hptlc.
Hptlc.Hptlc.
Hptlc.
 
opthalmic preparations, Classification,factors affecting for the drug given t...
opthalmic preparations, Classification,factors affecting for the drug given t...opthalmic preparations, Classification,factors affecting for the drug given t...
opthalmic preparations, Classification,factors affecting for the drug given t...
 
Scale up techniques in the production of sterile products &amp; ophthalmic pr...
Scale up techniques in the production of sterile products &amp; ophthalmic pr...Scale up techniques in the production of sterile products &amp; ophthalmic pr...
Scale up techniques in the production of sterile products &amp; ophthalmic pr...
 
Abdos Cataloge International
Abdos Cataloge InternationalAbdos Cataloge International
Abdos Cataloge International
 
Compendial methods for evalution of crude drug and herbal formulation
Compendial methods for evalution of crude drug and herbal formulationCompendial methods for evalution of crude drug and herbal formulation
Compendial methods for evalution of crude drug and herbal formulation
 
Tlc for pesticide analysis copy
Tlc for pesticide analysis copyTlc for pesticide analysis copy
Tlc for pesticide analysis copy
 
Practice Guidelines Endocopic disinfection & Reprocessing
Practice Guidelines Endocopic disinfection & ReprocessingPractice Guidelines Endocopic disinfection & Reprocessing
Practice Guidelines Endocopic disinfection & Reprocessing
 
Chemical Oxygen Demand
Chemical Oxygen DemandChemical Oxygen Demand
Chemical Oxygen Demand
 
Chemical Methods for Sterilization
Chemical Methods for  SterilizationChemical Methods for  Sterilization
Chemical Methods for Sterilization
 
Parenteral by SV Deshmane
Parenteral by SV DeshmaneParenteral by SV Deshmane
Parenteral by SV Deshmane
 
014.irrigation and intracanal medicaments
014.irrigation and intracanal  medicaments014.irrigation and intracanal  medicaments
014.irrigation and intracanal medicaments
 
Miscellaneous test .pdf
Miscellaneous test .pdfMiscellaneous test .pdf
Miscellaneous test .pdf
 
Medicall genetics lab manual
Medicall genetics lab manualMedicall genetics lab manual
Medicall genetics lab manual
 
Medicall genetics lab manual وراثة عملي
Medicall genetics lab manual وراثة عملي  Medicall genetics lab manual وراثة عملي
Medicall genetics lab manual وراثة عملي
 
in process quality control test for ophthalmic and parenterals
in process quality control test for ophthalmic and parenterals in process quality control test for ophthalmic and parenterals
in process quality control test for ophthalmic and parenterals
 
manufacturing of Parenterals.pdf
manufacturing of Parenterals.pdfmanufacturing of Parenterals.pdf
manufacturing of Parenterals.pdf
 
Human orosomucoid-2-elisa-kit
Human orosomucoid-2-elisa-kitHuman orosomucoid-2-elisa-kit
Human orosomucoid-2-elisa-kit
 
Cleaning of ot, lab
Cleaning of ot, labCleaning of ot, lab
Cleaning of ot, lab
 
Good Chromatographic Practices
Good Chromatographic PracticesGood Chromatographic Practices
Good Chromatographic Practices
 

Recently uploaded

Gurgaon iffco chowk 🔝 Call Girls Service 🔝 ( 8264348440 ) unlimited hard sex ...
Gurgaon iffco chowk 🔝 Call Girls Service 🔝 ( 8264348440 ) unlimited hard sex ...Gurgaon iffco chowk 🔝 Call Girls Service 🔝 ( 8264348440 ) unlimited hard sex ...
Gurgaon iffco chowk 🔝 Call Girls Service 🔝 ( 8264348440 ) unlimited hard sex ...soniya singh
 
Call Girls Madhapur 7001305949 all area service COD available Any Time
Call Girls Madhapur 7001305949 all area service COD available Any TimeCall Girls Madhapur 7001305949 all area service COD available Any Time
Call Girls Madhapur 7001305949 all area service COD available Any Timedelhimodelshub1
 
Call Girls Hyderabad Kirti 9907093804 Independent Escort Service Hyderabad
Call Girls Hyderabad Kirti 9907093804 Independent Escort Service HyderabadCall Girls Hyderabad Kirti 9907093804 Independent Escort Service Hyderabad
Call Girls Hyderabad Kirti 9907093804 Independent Escort Service Hyderabaddelhimodelshub1
 
Models Call Girls Electronic City | 7001305949 At Low Cost Cash Payment Booking
Models Call Girls Electronic City | 7001305949 At Low Cost Cash Payment BookingModels Call Girls Electronic City | 7001305949 At Low Cost Cash Payment Booking
Models Call Girls Electronic City | 7001305949 At Low Cost Cash Payment Bookingnarwatsonia7
 
Call Girls Gurgaon Parul 9711199012 Independent Escort Service Gurgaon
Call Girls Gurgaon Parul 9711199012 Independent Escort Service GurgaonCall Girls Gurgaon Parul 9711199012 Independent Escort Service Gurgaon
Call Girls Gurgaon Parul 9711199012 Independent Escort Service GurgaonCall Girls Service Gurgaon
 
hyderabad call girl.pdfRussian Call Girls in Hyderabad Amrita 9907093804 Inde...
hyderabad call girl.pdfRussian Call Girls in Hyderabad Amrita 9907093804 Inde...hyderabad call girl.pdfRussian Call Girls in Hyderabad Amrita 9907093804 Inde...
hyderabad call girl.pdfRussian Call Girls in Hyderabad Amrita 9907093804 Inde...delhimodelshub1
 
Russian Call Girls in Goa Samaira 7001305949 Independent Escort Service Goa
Russian Call Girls in Goa Samaira 7001305949 Independent Escort Service GoaRussian Call Girls in Goa Samaira 7001305949 Independent Escort Service Goa
Russian Call Girls in Goa Samaira 7001305949 Independent Escort Service Goanarwatsonia7
 
College Call Girls Hyderabad Sakshi 9907093804 Independent Escort Service Hyd...
College Call Girls Hyderabad Sakshi 9907093804 Independent Escort Service Hyd...College Call Girls Hyderabad Sakshi 9907093804 Independent Escort Service Hyd...
College Call Girls Hyderabad Sakshi 9907093804 Independent Escort Service Hyd...delhimodelshub1
 
Call Girls Kukatpally 7001305949 all area service COD available Any Time
Call Girls Kukatpally 7001305949 all area service COD available Any TimeCall Girls Kukatpally 7001305949 all area service COD available Any Time
Call Girls Kukatpally 7001305949 all area service COD available Any Timedelhimodelshub1
 
Hi,Fi Call Girl In Marathahalli - 7001305949 with real photos and phone numbers
Hi,Fi Call Girl In Marathahalli - 7001305949 with real photos and phone numbersHi,Fi Call Girl In Marathahalli - 7001305949 with real photos and phone numbers
Hi,Fi Call Girl In Marathahalli - 7001305949 with real photos and phone numbersnarwatsonia7
 
VIP Call Girls Hyderabad Megha 9907093804 Independent Escort Service Hyderabad
VIP Call Girls Hyderabad Megha 9907093804 Independent Escort Service HyderabadVIP Call Girls Hyderabad Megha 9907093804 Independent Escort Service Hyderabad
VIP Call Girls Hyderabad Megha 9907093804 Independent Escort Service Hyderabaddelhimodelshub1
 
Low Rate Call Girls In Bommanahalli Just Call 7001305949
Low Rate Call Girls In Bommanahalli Just Call 7001305949Low Rate Call Girls In Bommanahalli Just Call 7001305949
Low Rate Call Girls In Bommanahalli Just Call 7001305949ps5894268
 
Kukatpally Call Girls Services 9907093804 High Class Babes Here Call Now
Kukatpally Call Girls Services 9907093804 High Class Babes Here Call NowKukatpally Call Girls Services 9907093804 High Class Babes Here Call Now
Kukatpally Call Girls Services 9907093804 High Class Babes Here Call NowHyderabad Call Girls Services
 
2025 Inpatient Prospective Payment System (IPPS) Proposed Rule
2025 Inpatient Prospective Payment System (IPPS) Proposed Rule2025 Inpatient Prospective Payment System (IPPS) Proposed Rule
2025 Inpatient Prospective Payment System (IPPS) Proposed RuleShelby Lewis
 
Call Girls in Mohali Surbhi ❤️🍑 9907093804 👄🫦 Independent Escort Service Mohali
Call Girls in Mohali Surbhi ❤️🍑 9907093804 👄🫦 Independent Escort Service MohaliCall Girls in Mohali Surbhi ❤️🍑 9907093804 👄🫦 Independent Escort Service Mohali
Call Girls in Mohali Surbhi ❤️🍑 9907093804 👄🫦 Independent Escort Service MohaliHigh Profile Call Girls Chandigarh Aarushi
 

Recently uploaded (20)

Gurgaon iffco chowk 🔝 Call Girls Service 🔝 ( 8264348440 ) unlimited hard sex ...
Gurgaon iffco chowk 🔝 Call Girls Service 🔝 ( 8264348440 ) unlimited hard sex ...Gurgaon iffco chowk 🔝 Call Girls Service 🔝 ( 8264348440 ) unlimited hard sex ...
Gurgaon iffco chowk 🔝 Call Girls Service 🔝 ( 8264348440 ) unlimited hard sex ...
 
Call Girls Madhapur 7001305949 all area service COD available Any Time
Call Girls Madhapur 7001305949 all area service COD available Any TimeCall Girls Madhapur 7001305949 all area service COD available Any Time
Call Girls Madhapur 7001305949 all area service COD available Any Time
 
Call Girls Hyderabad Kirti 9907093804 Independent Escort Service Hyderabad
Call Girls Hyderabad Kirti 9907093804 Independent Escort Service HyderabadCall Girls Hyderabad Kirti 9907093804 Independent Escort Service Hyderabad
Call Girls Hyderabad Kirti 9907093804 Independent Escort Service Hyderabad
 
Models Call Girls Electronic City | 7001305949 At Low Cost Cash Payment Booking
Models Call Girls Electronic City | 7001305949 At Low Cost Cash Payment BookingModels Call Girls Electronic City | 7001305949 At Low Cost Cash Payment Booking
Models Call Girls Electronic City | 7001305949 At Low Cost Cash Payment Booking
 
Call Girls Gurgaon Parul 9711199012 Independent Escort Service Gurgaon
Call Girls Gurgaon Parul 9711199012 Independent Escort Service GurgaonCall Girls Gurgaon Parul 9711199012 Independent Escort Service Gurgaon
Call Girls Gurgaon Parul 9711199012 Independent Escort Service Gurgaon
 
hyderabad call girl.pdfRussian Call Girls in Hyderabad Amrita 9907093804 Inde...
hyderabad call girl.pdfRussian Call Girls in Hyderabad Amrita 9907093804 Inde...hyderabad call girl.pdfRussian Call Girls in Hyderabad Amrita 9907093804 Inde...
hyderabad call girl.pdfRussian Call Girls in Hyderabad Amrita 9907093804 Inde...
 
VIP Call Girls Lucknow Isha 🔝 9719455033 🔝 🎶 Independent Escort Service Lucknow
VIP Call Girls Lucknow Isha 🔝 9719455033 🔝 🎶 Independent Escort Service LucknowVIP Call Girls Lucknow Isha 🔝 9719455033 🔝 🎶 Independent Escort Service Lucknow
VIP Call Girls Lucknow Isha 🔝 9719455033 🔝 🎶 Independent Escort Service Lucknow
 
Russian Call Girls in Goa Samaira 7001305949 Independent Escort Service Goa
Russian Call Girls in Goa Samaira 7001305949 Independent Escort Service GoaRussian Call Girls in Goa Samaira 7001305949 Independent Escort Service Goa
Russian Call Girls in Goa Samaira 7001305949 Independent Escort Service Goa
 
Russian Call Girls South Delhi 9711199171 discount on your booking
Russian Call Girls South Delhi 9711199171 discount on your bookingRussian Call Girls South Delhi 9711199171 discount on your booking
Russian Call Girls South Delhi 9711199171 discount on your booking
 
Call Girl Dehradun Aashi 🔝 7001305949 🔝 💃 Independent Escort Service Dehradun
Call Girl Dehradun Aashi 🔝 7001305949 🔝 💃 Independent Escort Service DehradunCall Girl Dehradun Aashi 🔝 7001305949 🔝 💃 Independent Escort Service Dehradun
Call Girl Dehradun Aashi 🔝 7001305949 🔝 💃 Independent Escort Service Dehradun
 
College Call Girls Hyderabad Sakshi 9907093804 Independent Escort Service Hyd...
College Call Girls Hyderabad Sakshi 9907093804 Independent Escort Service Hyd...College Call Girls Hyderabad Sakshi 9907093804 Independent Escort Service Hyd...
College Call Girls Hyderabad Sakshi 9907093804 Independent Escort Service Hyd...
 
Call Girls Kukatpally 7001305949 all area service COD available Any Time
Call Girls Kukatpally 7001305949 all area service COD available Any TimeCall Girls Kukatpally 7001305949 all area service COD available Any Time
Call Girls Kukatpally 7001305949 all area service COD available Any Time
 
Hi,Fi Call Girl In Marathahalli - 7001305949 with real photos and phone numbers
Hi,Fi Call Girl In Marathahalli - 7001305949 with real photos and phone numbersHi,Fi Call Girl In Marathahalli - 7001305949 with real photos and phone numbers
Hi,Fi Call Girl In Marathahalli - 7001305949 with real photos and phone numbers
 
VIP Call Girls Hyderabad Megha 9907093804 Independent Escort Service Hyderabad
VIP Call Girls Hyderabad Megha 9907093804 Independent Escort Service HyderabadVIP Call Girls Hyderabad Megha 9907093804 Independent Escort Service Hyderabad
VIP Call Girls Hyderabad Megha 9907093804 Independent Escort Service Hyderabad
 
Low Rate Call Girls In Bommanahalli Just Call 7001305949
Low Rate Call Girls In Bommanahalli Just Call 7001305949Low Rate Call Girls In Bommanahalli Just Call 7001305949
Low Rate Call Girls In Bommanahalli Just Call 7001305949
 
Kukatpally Call Girls Services 9907093804 High Class Babes Here Call Now
Kukatpally Call Girls Services 9907093804 High Class Babes Here Call NowKukatpally Call Girls Services 9907093804 High Class Babes Here Call Now
Kukatpally Call Girls Services 9907093804 High Class Babes Here Call Now
 
Call Girls in Lucknow Esha 🔝 8923113531 🔝 🎶 Independent Escort Service Lucknow
Call Girls in Lucknow Esha 🔝 8923113531  🔝 🎶 Independent Escort Service LucknowCall Girls in Lucknow Esha 🔝 8923113531  🔝 🎶 Independent Escort Service Lucknow
Call Girls in Lucknow Esha 🔝 8923113531 🔝 🎶 Independent Escort Service Lucknow
 
2025 Inpatient Prospective Payment System (IPPS) Proposed Rule
2025 Inpatient Prospective Payment System (IPPS) Proposed Rule2025 Inpatient Prospective Payment System (IPPS) Proposed Rule
2025 Inpatient Prospective Payment System (IPPS) Proposed Rule
 
Call Girl Guwahati Aashi 👉 7001305949 👈 🔝 Independent Escort Service Guwahati
Call Girl Guwahati Aashi 👉 7001305949 👈 🔝 Independent Escort Service GuwahatiCall Girl Guwahati Aashi 👉 7001305949 👈 🔝 Independent Escort Service Guwahati
Call Girl Guwahati Aashi 👉 7001305949 👈 🔝 Independent Escort Service Guwahati
 
Call Girls in Mohali Surbhi ❤️🍑 9907093804 👄🫦 Independent Escort Service Mohali
Call Girls in Mohali Surbhi ❤️🍑 9907093804 👄🫦 Independent Escort Service MohaliCall Girls in Mohali Surbhi ❤️🍑 9907093804 👄🫦 Independent Escort Service Mohali
Call Girls in Mohali Surbhi ❤️🍑 9907093804 👄🫦 Independent Escort Service Mohali
 

Discussion of dialyzer reuse

  • 1. ● DISCUSSION ofDialyzer Reuse INTRODUCTION Dialysis facilities that reuse dialyzers must follow strict guidelines set forth by the AAMI (Association for the Advancement of Medical Instrumentation) The AAMI guidelines make provisions for the patient’s safety, as listed below: ● A dialyzer must be clearly labelled with the patient’s name and should only be used for the same patient. ● A dialyzer must be tested after each use to make sure it is working properly. ● A dialyzer must be tested after rinsing for any traces of disinfectant that may remain. Patients must be monitored for any reactions due to reuse. REQUIREMENTSFORREPROCESSINGDIALYZERS WATER ● This should be of AAMI standard or preferably of ultrapure quality. ● The water line pressure should be 13 kg/cm2 (20 psi), and every washing area should be equipped with 2 outlets or a T connection. ● Two different fittings will be required on the water line at each reprocessing area, a standard tubing to clean the blood compartment, and a Hansen connector for backwashing the dialysate compartment. ● Fittings should be of 316 SS or medical grade PVC only A prolonged rinse with Reverse Osmosis (RO) water followed by reverse ultra- filtration has been shown to be an effective means of cleaning dialyzers without the addition of chemicals CLEANING AND DISINFECTING AGENTS ● While selecting disinfectant we need to see for – Elimination of organic pollution, Inorganic component removal capacity, Disinfection efficacy, Safety and Environmental pollution. These agents should preferably be available online in the reprocessing areas. ● Overhead tanks containing the chemicals may be of 25 to 50 liters capacity and should be 
 refilled with fresh solutions every week, after cleaning. ● All tanks and piping for sodium hypochlorite should be composed of medical grade PVC, and those for formaldehyde, glutaraldehyde, and peracetic acid
  • 2. should be composed of 316SS. ● As per Indian Haemodialysis guidelines, common cleansing/disinfecting agents in use are mentioned below but regardless of the type of cleaning or disinfecting agent used in the facility, a thorough water rinse must be done prior to adding chemicals and after cleaning and disinfecting. Safety tests must be done after the final rinse to confirm the absence of any chemical used. ● It is dangerous to rely on a timed rinse without the use of a valid safety test. Residual testing prevents patient injury due to chemical exposure. a) Sodium Hypochlorite1–2% ● Sodium hypochlorite (bleach) is a cold disinfectant. Commercially available cans are 10% and they need to be diluted to the above concentration. ● There is formation of free chlorine that is a strong oxidant. It cleans and eliminates cellular debris. ● The advantages are that it is cheap, effective and safe. Residual testing for sodium hypochlorite is done with chlorine reagent test strips with a limit of 0.5 parts per million (ppm). It is a very sensitive method. Sodium hypochlorite in minute amounts greater than 1:25,000 produces hemolysis. ● Failure to perform residual testing will result in acute hemolysis or slow hemolysis that may go undetected. ● Sodium hypochlorite (bleach), diluted to 0.6% or less, dissolves proteinaceous deposits that may occlude fibers. Bleach increases albumin losses in high-flux cellulose triacetate (CT 190) and polysulfone. ● Polyvinylpyrrolidone dialyzers Albumin losses are generally not clinically significant unless bleach is used on high-flux membranes with exceptionally high water permeability. b) Hydrogen Peroxide ● Used as a cleaning agent. ● 1% or less concentration. ● It does not remove protein deposits.
  • 3. ● RO water should be used for dilution. ● Must be stored in airtight container as it is unstable. ● This is instilled in the dialysate compartment of the dialyzer rather than the blood compartment. ● Though hydrogen peroxide has been shown to have a lower potency than hypochlorite, 2 Indian studies showed good results after reuse with this agent. ● Test strips detect peroxide residues with limit of 0.5mg/l. ● Range: 0-0.5-2.5-5-10-25mg/l H2O2. ● Colour change: white > blue. ● Residues should be NIL. c) Formaldehyde (4%) ● It is the most common disinfectant used for fluid delivery systems. ● Formaldehyde has no cleaning properties. Formaldehyde denatures protein and fixes most cellular debris. ● Formaldehyde is a cold sterilant that effectively kills all microorganisms, including spores and resistant viruses but with risk of serious septicemia, if inadequate formaldehyde concentrations are being used. ● Commercially available as 40%, this can be diluted with the water used for reprocessing to give a final strength of 4%. ● Concentrations lower than 4% formaldehyde do not adequately kill Mycobacterium chelonae in water ● It is an inexpensive and stable solution with a long shelf-life. Formaldehyde is a gas that is dissolved in water to form the compound formalin. ● The formaldehyde gas is irritating to the eyes and has an offensive odour. Gloves must always be worn when handling formaldehyde to prevent dermatitis and allergic sensitivities. The room must be well ventilated. Any splashing must be minimized. A face shield gives total protection to the face. Minimally, eye protection (goggles) must be worn when handling formaldehyde.
  • 4. ● Sensitive residual testing is now available with limit of 1.0 ppm. The same principles apply as described for peracetic acid. In the past, safety tests for formaldehyde used Schiff’s reagent. Schiff’s reagent will test to 5 ppm. The newer indicator test strips are more sensitive. d) Glutaraldehyde (2%) ● This has to be freshly prepared and activated. ● The chemical potency of the solution may be tested with Schiff’s reagent which 
 produces a magenta colour similar to that seen with formaldehyde. ● Glutaraldehyde should be used for preserving the end caps, universal connectors, O- rings, and dialyzer caps when not in use. ● The solution should be replaced at intervals of not less than 10 days. ● Small containers containing glutaraldehyde should be available both at the dialysis 
 stations and at the reprocessing areas. e) PeraceticAcid(Renalin/Hemoclean, etc.) Hemoclean/Renalin ● A mixture of peracetic acid, hydrogen peroxide, & acetic acid ○ PERACETIC ACID: Disinfection during short term. ○ HYDROGEN PEROXIDE: Disinfection during long terms, Isolate protein pollutants. ○ ACETIC ACID: Cleaning effect (Remove inorganic pollutants) Strong bactericidal, sporicidal, fungicidal, & virucidal. ● It is probably the cold sterilant of choice for fluid delivery systems. Unlike formaldehyde, this mixture leaves no toxic residues. It decomposes into oxygen and acetic acid after reacting with organic material. ● The mixture is a strong oxidant that readily cleans all cellular debris, precipitates, and scale in the machines when used routinely. ● Minimum concentration 1%. It can be used as cleaning agent & sterilant. The undiluted solution should be diluted to prepare 2 solutions of the following: ● 2% (200 ml in 10 liters of water) as a cleaning agent and
  • 5. 3.5% (350 ml in 10 liters of water) as a disinfectant ● Minimum dwell time 11 hours. ● Stable for 1 week. ● Avoid direct sunlight exposure of solution. ● The odour is pungent. ● Test strips are available to check for the presence or absence of the peracetic acid mixture. Indicator test strips ensure the adequate presence of this cold sterilant in the machines and residual test strips test confirms absence. ● Range of indicator test strips: 0-500-1000-1500-2000 mg/l peracetic acid. ● Colour change: bright yellow > red. If the sterilant inside the dialyser shows less than 500 PPM it should not be used ideally as it suggests there is inadequate strilent effect. Now some other Countries specially Japan has increased this limit to 800 PPM. ● Once the confirmation is done that the potency is more than 500 or 800 PPM then prime the dialyser with NS. ● After the peracetic acid mixture is rinsed from the machine, residual test strips test for the absence of the peracetic acid mixture. The dialyzer should be primed with at least 2000 ml of 0.9% normal saline using the dialysis machine blood pump at a speed of 150 ml/ min. The dialysate lines should be connected and the dialysate compartment filled with dialysate flowing at 500 ml/ min prior to starting the priming procedure. Failure to “dialyze” the disinfectant out may result in inadequate removal and reactions after starting dialysis. ● After 2000 ml of saline priming, effluent from the venous line should be checked for the presence of residual disinfectant. ● The residual testing is very sensitive with limit <1 ppm. ● Because the peracetic acid mixture is a strong oxidant, its handling requires careful attention to avoid chemical burns. Gloves and face protection are mandatory. Accidental contact exposure to this chemical requires water flushes and medical attention similar to formaldehyde. REPROCESSING OF TUBINGS 
 ● No literature supports the reuse of blood tubings, however, they are reused in most units.
  • 6. ● The tubings are washed free of blood by treated water of AAMI or EU standard, and then with a 1.6% solution of sodium hypochlorite. ● The arterial and venous bubble chambers are gently tapped to release clots, and the side tubings are all cleared by clamping the outlets to dislodge any adherent material. ● The tubings are again rinsed with water and then connected to a supply of 4% formaldehyde, which is allowed to completely displace water and air from the tubings. TESTS OF PERFORMANCE ● No objective tests of performance are available for blood tubings. ● The tubings are discarded if the normal elasticity appears to be lost, if there are visible cracks, if there is a change from the normal transparent appearance, or damage to any of the hubs. ● A test that is sometimes useful is to compare the elasticity of the pump segment with that of new tubing on a blood pump. ● Failure to give 90% of the flow obtained with a new tubing segment or “slipping” of the tubing or a “slapping sound” from the rollers may indicate a malfunction of the pump segment of the tubing and require it to be discarded. Mention By- Anup Singh Dialysis Technician and Trainer 9651633399